How to setup Senorex Pressure Transducer
Daryl Hazelman
Posted on 22.08.2023 17:46
I have a FPG1DT,1AK,2U,5B,6A pressure transducer and need help in setting up the transducer for a IOLITE 6xSTG Module.
Daryl Hazelman
Posted on 23.08.2023 13:05
This is a sensortec transducer, not a sensorex. What do I set the Range and Bridge mode resistance to?
Rok Kmetič
Posted on 24.08.2023 12:59
Dear Daryl,
from the specifications that you have attached, it would be best that the Range is set to 20 mV/V. According to the specifications at max capacity, the raw value is around 5.3 mV/V.
But I am not sure about the Bridge resistance. It is not mentioned in the screenshots. You can measure the resistance by hand and input the measured value. Or send us the whole datasheet of the sensor.
The excitation voltage is correctly set to 10V.
Best regards.
Rok
signup
for
free
account
