Is it possible to connect the below pressure transmitter to MINITAURs?

https://www.wika.com.tr/upload/DS_PE8804_en_co_142...

Since it gives 0-10 Voltage, I guess it should be possible to do it and I guess I need to connect the transmitter via a 9 pin d'sub connector.

By the way the transmitter also has an option of giving 4-20 mA output signal. If I choose that option I think I'll need a DSI-20 mA adapter to connect it to MINITAURs. My other question is should I select the voltage or current option? Will it make any difference?

Cheers,

Volkan


Attached files:
DEWESoft Support
Technical support
Posted on 26.04.2018 14:45

Hi,

to connect the sensor to MinitaurS we recommend the voltage option.

Here is the connection that needs to be used:
Uv Pin 1 -> Exc+, Sns+
Out Pin 2 ->In+
Gnd Pin 3 ->In-, Exc-, Sns-

If a long cable will be used, it is recommended to connect IN- and GND near the sensor to avoid measuring the voltage drop.