Analogue output issues with SIRIUS

Emanuele Spinosa

Posted on 16.01.2019 16:00

Dear Dewesoft,


I am generating an analogue output from a SIRIUS. The shape of the output signal is determined by a function I define in a math channel.

However, I noticed that if I connect a voltage tester to any channel, even if I don't generate anything, I have an output of a few mV, which is undesired in my case.

In addition, this also happens sometimes when I generate a 0V output.


Do you know what can cause this? Thanks. 

Regards,


Emanuele

DEWESoft Support
Technical support
Posted on 18.01.2019 14:03

Dear Emanuele,


could you please provide a data file so we can see if this is a normal behaviour or it's something else causing this issue.


Regards



Emanuele Spinosa

Posted on 18.01.2019 14:22

Dear Dewesoft,


I will provide a data file, but the SIRIUS outputs about 20-30 mV even if the software is not running, I just saw this using a tester. Is this value within the uncertainty of the instrument?


Emanuele

DEWESoft Support
Technical support
Posted on 24.01.2019 15:29

Dear Emanuele,


we talked with our hardware developers, and this is their answer:


typical offset results at factory calibration is less than 5mV for 20kS/s and 80kS/s. Specified limit is 20mV. 

Only 2% of measurements that passed are above 5mV.


Values above specified limit 20mV are not within the uncertainty of the instrument. Please return the unit for check and calibration.


Regards