4-20mA Pressure Transducer Setup
Chris Warren
Posted on 07.09.2023 01:13
I am looking to set up a current output 4-20mA Omega PX309 pressure transducer. I started with the Ext. direct shunt using an external power supply and hooking up similar to the diagram shown below using the 2 and 7 pins. When doing this it was giving an overload error in the setup. Does anyone have any experience using the current output pressure transducers in dewesoft?
Rok Kmetič
Posted on 12.09.2023 10:36
Dear Chris,
the Krypton STG module is capable of providing the excitation for the sensor. You will have to use the External loop-powered shunt input type and connect the sensor as it is shown in the diagram. Excitation voltage can be set to 15 V because the sensor has declared excitation 9 to 30 V DC.
Another option is to use the DSI-20mA adapter which makes connecting the sensor easier without the need for soldering (it has a built-in shunt). More about the adapter you can find in the accessories manual on page 33: https://d36j349d8rqm96.cloudfront.net/3/6/Dewesoft-Accessories-and-sensors-Manual-EN.pdf
In your diagram, there is missing a shunt resistor. That is the reason you get the overload error.
Best regards.
Rok
Benjamin Solem
Posted on 14.10.2026 13:50
Hi Rok or other DEWESoft support team,
I'm looking into this setup with the 3STG Krypton module and DSI-20mA adapters for convenience for a Rosemount™ 3051 Pressure Transmitter. However, I wish to confirm that the current/power limit for the krypton is 45 mA (400 mW max. Power) per input channel? I assume so given the other information, but would like a confirmation. If I've done my math right the 15 V excitation would work in my case and stay below the 400 mW even at alarm levels (slightly above >20 mA).
A similar confirmation would be nice for the DSI adapter though. I assume the power rating of 250 mA relates to the shunt resistor and the input current, not the excitation side? I.e. the Krypton 400 mW is what limits the current draw on the excitation side.
BR,