RTD sensor wiring
I need to connect a generic 3 wire RTD sensor to a IOLITEi-1xSTG through a DSI-ACC adapter.
The IOLITE is powered by the computer through ethernet.
I followed the "Accessiories and sensors manual" which provides the following wiring schematics at page 21.
From what I understand, I need to connect only pins 2, 3, 4. The jumps between 1-2 and 3-4-5 are internal to the DAQ.
I solded the 5-pin BINDER connector following the 3 wire schema schema but on setup inside the DewesoftX software the 4 wire schema is shown and the sensor reads almost no resistance.
What am I doing wrong? Thank you in advance.
Dear Davide,
Thank you for reaching us. Unfortunately, the DSI-ACC adaptor is not suitable for resistance measurement. But I think this was just a typo because you are referring to page 21 in the accessories manual which is correct. Only the DSI-RTD adaptor has a built-in binder connector.
You have made the correct wiring, but if you want to use the 3-wire connection you need to manually connect pins 1 & 2 together and also the pins 3 & 5. This will give the RTD sensor power and the measurement will be correct.
DewesoftX has a ''universal'' picture for the measurement (with the dotted lines), for the manual, we have made all connections options more visible.
In the picture below I have marked the connections:
Please let me know if this would solve the problem.
Best Regards.
Rok
Thank you for the reply.
As you said it was a typo, sorry for that, I'm using a DSI-RTD adapter.
I connected the sensor as you showed in the picture but I'm encountering the same problems, I'll leave you two screenshots
My mistake due to inexperience! A little short circuit, thank you for the kind reply.
Dear Davide,
I'm glad to hear that you were able to sort it out and thanks for letting me know.
Best Regards.
Rok